Achieving Target Temperature

Hello,

I am testing a QSI 6120 that I recently acquired and am trying to collect calibration frames at -20 degrees. What is odd is that it seems to reach -18 degrees and SGP reports that as 100%. Is the acceptance zone described and programmed somewhere?

Thanks

Farzad

image

I believe that the 100% refers to the power level being set for the Thermo electric Cooler. The cooler has been set to 100% power and cannot go any higher. The -18C is as cool as that cooler can go under your ambient temperature. It does not have enough power to get to your set point of -22C.

So it is not that SGP is reporting it has 100% achieved your setpoint temperature, rather it is reporting that it is at 100% max power for the TEC and the best it can do in achieving your setpoint temperature is -18C.

100% means that the cooler power is at 100%. So with your ambient air temperature, -18.1C appears to be as low as the cooler can get the camera at 100% power.

Okay, thanks. I thought the percentage was a relation between the target temperature and current temperature.

Farzad